Int J Infect Dis. 2019 Jan 8. pii: S1201-9712(19)30005-0. doi: 10.1016/j.ijid.2018.12.014. [Epub ahead of print]
[h=1]Comparison of the Cepheid Xpert Xpress Flu/RSV assay and commercial real-time PCR for the detection of influenza A and influenza B in a prospective cohort from China.[/h] Zou X[SUP]1[/SUP], Chang K[SUP]1[/SUP], Wang Y[SUP]1[/SUP], Li M[SUP]1[/SUP], Zhang W[SUP]1[/SUP], Wang C[SUP]1[/SUP], Lu B[SUP]1[/SUP], Xiong Z[SUP]1[/SUP], Han J[SUP]1[/SUP], Zhang Y[SUP]1[/SUP], Zhao J[SUP]1[/SUP], Cao B[SUP]2[/SUP]; for CAP-China Network.
[h=3]Author information[/h]
[h=3]Abstract[/h] [h=4]BACKGROUND:[/h] The Xpert Xpress Flu/RSV assay is released by FDA for rapid detection of influenza A (FluA), influenza B (FluB), and respiratory syncytial virus (RSV). This study aimed to evaluate its clinical performance in comparison to that of the RT-PCR assay cleared by China FDA (CFDA-PCR).
[h=4]METHODS:[/h] Nasopharyngeal specimens were collected from patients and tested by the two assays side by side. Discordant results were tested with a laboratory-developed real-time PCR for resolution. Viral load in the sample was quantified with a droplet digital PCR.
[h=4]RESULTS:[/h] A total of 658 specimens were involved and gave 94.7% ∼ 99.1% agreement between the two assays. The Xpert assay showed higher sensitivity for FluA (100% vs. 89.8%) and FluB detection (100% vs. 95.3%), and also higher accuracy (98.9% vs. 95.7%) for FluA than the CDFA-PCR. The positive and negative predictive values (NPV) for the three viruses ranged from 90.5% to 100% in the two assays, with higher NPV for FluA and FluB in Xpert assay. Moreover, the Xpert Ct values showed a linear correlation with virus titer in specimens tested.
[h=4]CONCLUSION:[/h] Overall, the Xpert assay is a reliable and sensitive tool for the detection of FluA, FluB and RSV in our clinical settings.
Copyright ? 2019. Published by Elsevier Ltd.
[h=4]KEYWORDS:[/h] RT-PCR; Xpert Xpress Flu/RSV; digital PCR; sensitivity
PMID: 30634045 DOI: 10.1016/j.ijid.2018.12.014
[h=1]Comparison of the Cepheid Xpert Xpress Flu/RSV assay and commercial real-time PCR for the detection of influenza A and influenza B in a prospective cohort from China.[/h] Zou X[SUP]1[/SUP], Chang K[SUP]1[/SUP], Wang Y[SUP]1[/SUP], Li M[SUP]1[/SUP], Zhang W[SUP]1[/SUP], Wang C[SUP]1[/SUP], Lu B[SUP]1[/SUP], Xiong Z[SUP]1[/SUP], Han J[SUP]1[/SUP], Zhang Y[SUP]1[/SUP], Zhao J[SUP]1[/SUP], Cao B[SUP]2[/SUP]; for CAP-China Network.
[h=3]Author information[/h]
[h=3]Abstract[/h] [h=4]BACKGROUND:[/h] The Xpert Xpress Flu/RSV assay is released by FDA for rapid detection of influenza A (FluA), influenza B (FluB), and respiratory syncytial virus (RSV). This study aimed to evaluate its clinical performance in comparison to that of the RT-PCR assay cleared by China FDA (CFDA-PCR).
[h=4]METHODS:[/h] Nasopharyngeal specimens were collected from patients and tested by the two assays side by side. Discordant results were tested with a laboratory-developed real-time PCR for resolution. Viral load in the sample was quantified with a droplet digital PCR.
[h=4]RESULTS:[/h] A total of 658 specimens were involved and gave 94.7% ∼ 99.1% agreement between the two assays. The Xpert assay showed higher sensitivity for FluA (100% vs. 89.8%) and FluB detection (100% vs. 95.3%), and also higher accuracy (98.9% vs. 95.7%) for FluA than the CDFA-PCR. The positive and negative predictive values (NPV) for the three viruses ranged from 90.5% to 100% in the two assays, with higher NPV for FluA and FluB in Xpert assay. Moreover, the Xpert Ct values showed a linear correlation with virus titer in specimens tested.
[h=4]CONCLUSION:[/h] Overall, the Xpert assay is a reliable and sensitive tool for the detection of FluA, FluB and RSV in our clinical settings.
Copyright ? 2019. Published by Elsevier Ltd.
[h=4]KEYWORDS:[/h] RT-PCR; Xpert Xpress Flu/RSV; digital PCR; sensitivity
PMID: 30634045 DOI: 10.1016/j.ijid.2018.12.014