• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

J Clin Virol Plus . Field performance evaluation of the PanBio rapid SARS-CoV-2 antigen assay in an epidemic driven by the B.1.351 variant in the E

tetano

Editor, Senior Moderator
J Clin Virol Plus


. 2021 Jun;1(1):100013.
doi: 10.1016/j.jcvp.2021.100013. Epub 2021 Apr 7.
Field performance evaluation of the PanBio rapid SARS-CoV-2 antigen assay in an epidemic driven by the B.1.351 variant in the Eastern Cape, South Africa


Oluwakemi Laguda Akingba[SUP] 1 2 [/SUP], Kaitlin Sprong[SUP] 1 [/SUP], Gert Marais[SUP] 1 3 [/SUP], Diana Ruth Hardie[SUP] 1 3 [/SUP]



Affiliations

Abstract

Background: South Africa was the African country with the most recorded cases of SARS-CoV-2 during 2020, experiencing 2 waves of infection. During the first wave, diagnostics were largely based on reverse transcription-linked PCR (RT-PCR). The Abbott PanBio antigen test was deployed during the 2nd wave which may have been driven by emergence of the B.1.351 variant. At the time of evaluation in mid-November 2020, B.1.351 was the dominant circulating virus in Nelson Mandela Bay, in the Eastern Cape Province.
Methods: Used PanBio antigen swabs (collected from patients with genetically characterised virus) were first validated as suitable for PCR. A prospective study was then undertaken to evaluate assay performance in the field. Testing was conducted at mobile community testing centres on 677 ambulant patients. Used swabs were kept and tested by RT-PCR.
Results: During initial validation, used swabs in proprietary lysis buffer were found to be suitable for PCR and secondly, the PB assay reliably detected patients infected with B.1.351. In the field study, of 146 RT-PCR positive individuals, 101 were RTD positive in the clinic. The RTD had a sensitivity of 69.2% (95%CI 61.4, 75.8) and specificity of 99.0% (95%CI 98.8, 99.3). Sensitivity was dependent on the amount of viral RNA in clinical samples, as reflected by the PCR cycle threshold (CT) value.
Conclusions: The assay reliably detected B.1.351 infections in ambulatory ill patients during initial validation and in field testing. In the field, assay sensitivity was >90% in patients with high viral loads who are expected to be most infectious. Negative and positive predictive values were also >90%.

Keywords: 501Y.v2; 501Y.v2, 501Y variant 2; B.1.351; COVID-19; COVID-19, Coronavirus disease 2019; CT, cycle threshold; GISAID, global initiative for sharing all influenza data; IQR, interquartile range; NP, nasopharyngeal; PCR, polymerase chain reaction; Point of care; RT-PCR, reverse transcription polymerase chain reaction; RTD, rapid test device; Rapid antigen test; SARS-CoV-2; SARS-CoV-2, severe acute respiratory syndrome coronavirus 2; WHO, world health organisation.
 