• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Analytical performance of the BD veritor? system for rapid detection of influenza virus A and B in a primary healthcare setting

tetano

Editor, Senior Moderator
BMC Infect Dis. 2016 Sep 9;16:481. doi: 10.1186/s12879-016-1811-9.
[h=1]Analytical performance of the BD veritor? system for rapid detection of influenza virus A and B in a primary healthcare setting.[/h] Mese S[SUP]1[/SUP], Akan H[SUP]2[/SUP], Badur S[SUP]3[/SUP], Uyanik A[SUP]3[/SUP]; Istanbul Rapid Test Study Group.
[h=3]Collaborators (10)[/h]

[h=3]Author information[/h]

[h=3]Abstract[/h] [h=4]BACKGROUND:[/h] Infections with influenza A virus cannot be clinically differentiated from infections caused by influenza B virus or other respiratory viruses. Additionally, although antiviral treatment is available for influenza A virus, it is not effective for the other viruses and must be initiated early in the course of disease for it to be effective. For these reasons, there is a need for a rapid, accurate diagnostic test for use in physicians' offices at the time patients are seen. We report the first field performance of BD Veritor? System for Rapid Detection of Flu A + B test compared to real time PCR. The performance of this test was compared to real time PCR performed in the Istanbul University Influenza Reference Laboratory.
[h=4]METHOD:[/h] A single-blinded cross sectional study was conducted in nine different family medicine centers in Istanbul, Turkey between 01 November 2014 and 01 May 2015. For every patient, two specimens were collected, one for real time PCR and one for the Veritor test. Specimens for the Veritor test were immediately tested at the participating clinic according to the manufacturer's instructions. The specimens for real time PCR were transferred to the reference laboratory.
[h=4]RESULTS:[/h] A total of 238 persons were included in the study: 72 (30 %) of the patients included in the study were below 19 years old and accepted as childhood group. Mean age of adults was 42.4 and children 10.2 years. A total of 122 patients out of 238 were positive for influenza. The clinical sensitivity and specificity of the Veritor test in all age groups was determined to be 80 and 94 %, respectively. Positive predictive value was 93 % and the negative one was 81 %.
[h=4]CONCLUSION:[/h] Field performance of the rapid influenza test was high and found to be useful with respect to rational antiviral use, avoiding unnecessary antibiotic usage and the management of cases by the family physicians.


[h=4]KEYWORDS:[/h] Point of Care; Rapid diagnostic test; Seasonal influenza; Veritor clinical sensitivity and specificity

PMID: 27612949 DOI: 10.1186/s12879-016-1811-9
[PubMed - in process] Free full text
 