• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Clin Chem Lab Med . Clinical assessment of SNIBE Maglumi SARS-CoV-2 antigen fully-automated chemiluminescent immunoassay

tetano

Editor, Senior Moderator
Clin Chem Lab Med


. 2023 Mar 8.
doi: 10.1515/cclm-2023-0181. Online ahead of print.
Clinical assessment of SNIBE Maglumi SARS-CoV-2 antigen fully-automated chemiluminescent immunoassay


Laura Pighi[SUP] 1 2 [/SUP], Brandon M Henry[SUP] 3 [/SUP], Simone De Nitto[SUP] 1 2 [/SUP], Gianluca Gianfilippi[SUP] 4 [/SUP], Gian Luca Salvagno[SUP] 1 2 [/SUP], Mario Plebani[SUP] 5 [/SUP], Giuseppe Lippi[SUP] 1 [/SUP]



Affiliations

Abstract

Objectives: Given that SARS-CoV-2 antigen tests will represent a pillar for supporting or surrogating molecular testing in the endemic period, we report here the clinical performance of the new SNIBE Maglumi SARS-CoV-2 antigen fully-automated chemiluminescent immunoassay (MAG-CLIA SARS-CoV-2 Ag).
Methods: The study population consisted of 181 subjects (mean age 61 ± 21 years; 92 females) undergoing coronavirus disease 2019 (COVID-19) testing at the local diagnostic facility, from December 2022 to February 2023. Routine diagnostic practice involved the collection of a double nostril nasopharyngeal swab, analyzed in duplicate with SARS-CoV-2 antigen (MAG-CLIA SARS-CoV-2 Ag) and molecular (Altona Diagnostics RealStar SARS-CoV-2 RT-PCR Kit) tests.
Results: A significant Spearman's correlation was found between MAG-CLIA SARS-CoV-2 Ag and mean Ct values of SARS-CoV-2 E and S genes (r=-0.95; p<0.001). In all nasopharyngeal samples, the area under the curve (AUC) of MAG-CLIA SARS-CoV-2 Ag was 0.86 (95% CI, 0.81-0.90), with 0.71 sensitivity and 1.00 specificity at 7 ng/L cut-off, increasing to 0.98 (95% CI, 0.96-1.00) AUC and 0.96 sensitivity (with 0.97 specificity) in high viral load samples. When SARS-CoV-2 N protein concentration was replaced with raw instrumental readings (i.e., relative light units [RLU]), the AUC in all samples increased to 0.94. A RLU value of 945 was associated with 88.4% accuracy, 0.85 sensitivity, 0.95 specificity, 0.77 negative predictive value (NPV) and 0.97 positive predictive value (PPV), respectively.
Conclusions: We found satisfactory analytical performance of MAG-CLIA SARS-CoV-2 Ag, which could be used as surrogate of molecular testing for identifying high viral load samples. Broadening the reportable range of values may generate even better performance.

Keywords: COVID-19; SARS-CoV-2; antigen; immunoassay; infection.
 