• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Diagnostic accuracy of the real-time PCR cobas? Liat? Influenza A/B assay and the Alere i Influenza A&B NEAR isothermal nucleic acid amplification ass

tetano

Editor, Senior Moderator
J Clin Virol. 2017 Jul 26;94:86-90. doi: 10.1016/j.jcv.2017.07.012. [Epub ahead of print]
[h=1]Diagnostic accuracy of the real-time PCR cobas? Liat? Influenza A/B assay and the Alere i Influenza A&B NEAR isothermal nucleic acid amplification assay for the detection of influenza using adult nasopharyngeal specimens.[/h] Young S[SUP]1[/SUP], Illescas P[SUP]2[/SUP], Nicasio J[SUP]2[/SUP], Sickler JJ[SUP]3[/SUP].
[h=3]Author information[/h]

[h=3]Abstract[/h] [h=4]BACKGROUND:[/h] Accurate detection of influenza requires diagnostic testing; however, methods such as RADTs and central laboratory-based tests are limited by low sensitivity and time constraints, respectively.
[h=4]OBJECTIVE:[/h] To compare the performances of the cobas[SUP]?[/SUP] Liat[SUP]?[/SUP] Influenza A/B and Alere? i Influenza A&B point-of-care (POC) assays for detecting influenza A and B viruses using fresh nasopharyngeal specimens with the GenMark Dx[SUP]?[/SUP] Respiratory Viral Panel as the reference method, a FDA cleared IVD PCR test.
[h=4]STUDY DESIGN:[/h] A total of 87 samples collected in viral transport medium from adults ≥18 years of age were re-tested on both POC assays (based on the reference PCR method, 29 were influenza A and 18 were influenza B virus positive).
[h=4]RESULTS:[/h] The overall sensitivity and specificity of the cobas Influenza A/B for the detection of influenza A and B relative to reference PCR was 97.9% (95% confidence interval [CI] 88.9%, 99.6%) and 97.5% (95% CI: 87.1%, 99.6%), respectively, while the sensitivity of the Alere i Influenza A&B assay relative to the reference PCR method was 63.8% (95% CI: 49.5%, 76.0%) and the specificity was 97.5% (95% CI: 87.1%, 99.6%). The individual sensitivities and specificities of the cobas Influenza A/B assay for influenza A alone and influenza B alone were comparable to those of the reference PCR method (influenza A: sensitivity of 100% [95% CI: 88.3%, 100.0%] and specificity of 98.3% [95% CI: 90.9%, 99.7%]; influenza B: sensitivity of 94.4% [95% CI: 74.2%, 99.0%] and specificity of 100% [95% CI: 94.7%, 100.0%]). For the Alere i Influenza A&B assay, the individual specificities for influenza A and B were comparable to those of the reference PCR method (98.3% [95% CI: 90.9%, 99.7%] and 97.1% [95% CI: 90.0%, 99.2%], respectively), while the individual sensitivities were low relative to reference PCR (55.2% [95% CI: 37.5%, 71.6%] and 72.2% [95% CI: 49.1%, 87.5%], respectively).
[h=4]CONCLUSION:[/h] The cobas Influenza A/B assay demonstrated performance equivalent to laboratory-based PCR, and could replace rapid antigen tests.
Copyright ? 2017 Elsevier B.V. All rights reserved.


[h=4]KEYWORDS:[/h] Alere i Influenza A&B assay; Influenza virus infection; Point-of-care; Real-time polymerase chain reaction; cobas(?) Liat(?) Influenza A/B assay

PMID: 28772170 DOI: 10.1016/j.jcv.2017.07.012
 