• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

J Clin Lab Anal . Comparison of next generation diagnostic systems (NGDS) for the detection of SARS-CoV-2

tetano

Editor, Senior Moderator
J Clin Lab Anal


. 2022 Feb 17;e24285.
doi: 10.1002/jcla.24285. Online ahead of print.
Comparison of next generation diagnostic systems (NGDS) for the detection of SARS-CoV-2


Antonio O Sanchez[SUP] 1 [/SUP], Anna R Ochoa[SUP] 1 2 [/SUP], Sallie L Hall[SUP] 1 [/SUP], Chet R Voelker[SUP] 1 [/SUP], Rachel E Mahoney[SUP] 1 2 [/SUP], Jennifer S McDaniel[SUP] 1 3 [/SUP], August Blackburn[SUP] 4 [/SUP], Susana N Asin[SUP] 1 [/SUP], Tony T Yuan[SUP] 4 [/SUP]



Affiliations

Abstract

Introduction: The World Health Organization (WHO) declared coronavirus disease 2019 (COVID-19) a pandemic in March 2020. Initially, supply chain disruptions and increased demand for testing led to shortages of critical laboratory reagents and inadequate testing capacity. Thus, alternative means of biosample collection and testing were essential to overcome these obstacles and reduce viral transmission. This study aimed to 1) compare the sensitivity and specificity of Cepheid GeneXpert[SUP]®[/SUP] IV and BioFire[SUP]®[/SUP] FilmArray[SUP]®[/SUP] 2.0 next generation detection systems to detect SARS-CoV-2, 2) evaluate the performance of both platforms using different biospecimen types, and 3) assess saline as an alternative to viral transport media (VTM) for sample collection.
Methods: A total of 1,080 specimens consisting of nasopharyngeal (NP) swabs in VTM, NP swabs in saline, nasal swabs, oropharyngeal (OP) swabs, and saliva were collected from 216 enrollees. Limit of detection (LoD) assays, NP VTM and NP saline concordance, and saliva testing were performed on the BioFire[SUP]®[/SUP] FilmArray[SUP]®[/SUP] 2.0 Respiratory Panel 2.1 and Cepheid GeneXpert[SUP]®[/SUP] Xpress SARS-CoV-2/Flu/RSV assays.
Results: LoD and comparative testing demonstrated increased sensitivity with the Cepheid compared with the BioFire[SUP]®[/SUP] in detecting SARS-CoV-2 in NP VTM and saline, nasal, and OP swabs. Conversely, saliva testing on the Cepheid showed statistically significant lower sensitivity compared to the BioFire[SUP]®[/SUP] . Finally, NP swabs in saline showed no significant difference compared with NP swabs in VTM on both platforms.
Conclusion: The Cepheid and BioFire[SUP]®[/SUP] NGDS are viable options to address a variety of public health needs providing rapid and reliable, point-of-care testing using a variety of clinical matrices.

Keywords: BioFire; COVID-19; Cepheid; SARS-CoV-2; biospecimen; next generation detection systems.
 