• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

J Med Virol . Clinical Performance of Roche cobas 6800, Luminex ARIES, MiRXES Fortitude Kit 2.1, altona RealStar, and Applied Biosystems TaqPath fo

tetano

Editor, Senior Moderator
J Med Virol


. 2021 Mar 14.
doi: 10.1002/jmv.26940. Online ahead of print.
Clinical Performance of Roche cobas 6800, Luminex ARIES, MiRXES Fortitude Kit 2.1, altona RealStar, and Applied Biosystems TaqPath for SARS-CoV-2 Detection in Nasopharyngeal Swabs


Chun Kiat Lee[SUP] 1 [/SUP], Jason Wei Ming Tham[SUP] 1 [/SUP], Siyu Png[SUP] 1 [/SUP], Chean Nee Chai[SUP] 1 [/SUP], Shu Chi Ng[SUP] 1 [/SUP], Eunice Jia Min Tan[SUP] 1 [/SUP], Li Jie Ng[SUP] 1 [/SUP], Rui Ping Chua[SUP] 1 [/SUP], Musa Sani[SUP] 1 [/SUP], Yiqi Seow[SUP] 2 [/SUP], Gabriel Yan[SUP] 3 [/SUP], Julian Tang[SUP] 4 [/SUP]



Affiliations

Abstract

Objectives: We compared the performance of five assays for SARS-CoV-2 detection on nasopharyngeal swab samples: Roche 'cobas', Luminex 'ARIES', MiRXES 'Fortitude', altona 'RealStar', and Thermo Fisher Scientific 'TaqPath'.
Methods: A total of 94 nasopharyngeal swab samples were obtained from 80 confirmed COVID-19 cases in the first 2 weeks of illness (median: 7 days; range: 2-14 days) and 14 healthy controls. After collection, all samples were transported to the hospital clinical laboratory within 24 hours. These samples were tested on all five assays within three days of sample receipt.
Results: Of the 94 samples, 69 yielded the same result on all platforms, resulting in an agreement of 73.4% (69/94). Of these, 14 were the healthy control swabs which all tested negative, demonstrating good specificity across all platforms. The ARIES assay had the lowest detection rate (68.8%), followed by Fortitude (85.0%), RealStar (86.3%), cobas (95.0%), and TaqPath (100%). Statistically significant differences were observed for ARIES, Fortitude, and RealStar when compared against the best performing TaqPath using the McNemar's chi-squared test. A consensus result was established based on the results obtained by the cobas, Fortitude, RealStar, and TaqPath. Six discrepancies had failed to reach a consensus and were adjudicated using the Cepheid Xpert Xpress SARS-CoV-2.
Conclusion: Overall, the TaqPath and cobas assays were the most sensitive at detecting their designated SARS-CoV-2 gene targets. On the other hand, the ARIES assay was the least sensitive, thus warranting the need for assay re-optimisation prior to go-live at the testing laboratory. This article is protected by copyright. All rights reserved.

Keywords: COVID-19; SARS-CoV-2; coronavirus; molecular diagnostics; pandemic.
 